laboratory communications
Measurement of the intrinsic variability within protein crystals: implications for sample-evaluation and data-collection strategies
aDepartment of Physics, University of Oxford, Keble Road, Oxford OX1 3RH, England, bEuropean Molecular Biology Laboratory, Grenoble Outstation, 6 Rue Jules Horowitz, 38042 Grenoble, France, and cUnit of Virus Host-Cell Interactions, Université Grenoble Alpes–EMBL–CNRS, 6 Rue Jules Horowitz, 38042 Grenoble, France
*Correspondence e-mail: mbowler@embl.fr
The advent of micro-focused X-ray beams has led to the development of a number of advanced methods of sample evaluation and data collection. In particular, multiple-position data-collection and helical oscillation strategies are now becoming commonplace in order to alleviate the problems associated with radiation damage. However, intra-crystal and inter-crystal variation means that it is not always obvious on which crystals or on which region or regions of a crystal these protocols should be performed. For the automation of this process for large-scale screening, and to provide an indication of the best strategy for data collection, a metric of crystal variability could be useful. Here, measures of the intrinsic variability within protein crystals are presented and their implications for optimal data-collection strategies are discussed.
Keywords: microfocus; automation; sample evaluation; macromolecular crystallography.
1. Introduction
In the last ten years, protein micro-crystallography (Cusack et al., 1998; Riekel et al., 2005) has moved from a highly unusual and specialized technique to a standard method in modern structural biology (Smith et al., 2012). The large multi-component complexes and membrane proteins now routinely studied tend to produce either very small crystals or crystals that can be extremely heterogeneous in their diffraction properties. The increasing availability of micro-focused or micro X-ray beams (diameter < 10 µm; Flot et al., 2010; Sanishvili et al., 2008; Axford et al., 2012) with experimental environments optimized for macromolecular crystallography has led to the evolution of advanced sample-evaluation and data-collection protocols, such as mesh scans and helical oscillations (Bowler et al., 2010; Cherezov et al., 2009; Flot et al., 2010; Song et al., 2007; Aishima et al., 2010; Hilgart et al., 2011). The ability of a microbeam to spread the effect of radiation damage across a larger volume of a crystal or to resolve the most ordered part of a crystal that is relatively larger than the beam size (Sanishvili et al., 2008) has brought a much larger variety of samples to these beamlines (Flot et al., 2010). In many of the most challenging projects it has proved essential, rather than advantageous, to be able to collect data from multiple positions or to use helical oscillations (Efremov et al., 2010; Warne et al., 2008). Recently, a number of automated procedures to define the best position within a crystal from which to collect data have been described (Bowler et al., 2010; Sanishvili et al., 2008; Song et al., 2007; Aishima et al., 2010; Hilgart et al., 2011). However, a measure of the variability of diffraction quality within a crystal would provide an indication of the degree of both within and between crystals of a certain macromolecule and could act as an initial guide to beamline choice and data-collection strategy. Here, metrics of the innate variability within crystals are presented. A number of systems, varying from large multi-component complexes to membrane proteins and small soluble proteins, have been tested. The values provide an initial measure of the extent of variability and could be used to indicate whether a large or small beam should be used and whether single-position or multiple-position strategies should be employed. As the level of automation and the number of samples requiring evaluation increases at synchrotron facilities in the future, automated pre-evaluation and measurement of variability will be essential in the selection of optimized data-collection facilities and strategies.
2. Experimental procedures and results
2.1. Crystal preparation
Crystals of bovine mitochondrial F1-ATPase were grown as described previously (Lutter et al., 1993), with the exception that azide and ADP were omitted from all buffers. The crystals were mounted on MicroMesh loops (MiTeGen, Ithaca, New York, USA) and conditioned with the HC1b humidity-control device (Sanchez-Weatherby et al., 2009; Russi et al., 2011; Wheeler et al., 2012), using conditions previously established for maximum improvement of crystal diffraction properties (Bowler et al., 2006).
Recombinant human RhoA crystals were grown, harvested on a MicroMesh (MiteGen, Ithaca, New York, USA) and cryocooled directly as described previously (Pellegrini et al., 2011). Crystals of the closed conformation of β-phosphoglucomutase (β-PGM) from Lactobacillus lactis were grown as described previously (Baxter et al., 2010; Griffin et al., 2012). Crystals of thermolysin from Bacillus thermoproteolyticus were grown according to established protocols (Mueller-Dieckmann et al., 2005). Crystals of human PGK were grown as described previously (Cliff et al., 2010). Ultralente insulin (Ultratard, Novo Nordisk, Denmark) and glucose isomerase micro-crystals (Sigma–Aldrich) were swept out of solution using a MicroMesh (MiTeGen, Ithaca, New York, USA), excess liquid was removed and they were cryocooled directly (Pellegrini et al., 2011).
All other crystals were provided by their owners for testing; particular thanks are due to Tony Warne and Venki Ramakrishnan (MRC Laboratory of Molecular Biology, Cambridge, England) for crystals of β1-andregenic G-protein coupled receptor from turkey, prepared as described in Warne et al. (2008), and of the 70S ribosome from Thermus thermophilus, prepared as described in Selmer et al. (2006), respectively.
2.2. Mesh scans, line scans and data processing
Mesh scans were launched as described in Bowler et al. (2010) using the workflow interface in the MXCuBE beamline GUI (Brockhauser et al., 2012; Gabadinho et al., 2010). Briefly, a mesh is defined by drawing a box over the area of interest on the image from the beamline video microscope. The number of vertical and horizontal steps is then defined to produce a mesh where an image will be taken at the intersection of lines; here, the step size was similar to the beam diameter in order avoid overlapping areas. These points are converted to positions of the goniometer centring table and translation motors, and data collection is then performed at the corresponding positions. Line scans are similar to mesh scans except that only two points are defined and the number of images to be collected between these points (the step size) is defined. Line scans were used when assessing the diffraction quality along needle-shaped crystals. Diffraction images were processed using an EDNA (Incardona et al., 2009) characterization plugin running labelit.distil (Sauter et al., 2004; Zhang et al., 2006), MOSFLM (Powell et al., 2013) and BEST (Bourenkov & Popov, 2010), producing a large number of metrics for the comparison of diffraction quality. The choice of metric for diffraction quality is not always obvious. Here, we have selected the total integrated signal above background (TIS) as the best way to measure variation within a crystal, as it provides a good measure of the differences in diffraction intensities. Other metrics, such as the number of spots, are also a good measure of the quality, but as the higher resolution spots are often missed, differences between positions may not be highlighted. Calculation of differences using the number of spots leads to a similar trend as for the total signal above background, but is less discriminating (data not shown). The calculations presented here are applicable to any measure of quality. The diameter of physically contiguous images with TIS values within 10% of each other was also output in order to give an indication of the area of the best regions. Images without diffraction, those containing ice rings or weak diffraction from `glancing blows' were excluded from variability calculations by only including images with a number of counts above a threshold value.
2.3. Measuring intra-crystal variability
What is the best strategy for data collection? If a crystal diffracts homogenously then the best option is to match the beam size to the crystal and use the full diffraction power of the crystal and distribute the dose across a larger sample volume. If the quality of the crystal varies then the best strategy will be to use only the most ordered volumes. In this study, the diffraction characteristics of a large variety of samples has been probed with X-ray beams of varying size (Table 1), including a number of challenging projects where particular data-collection strategies using either large (≥100 µm diameter) or small (≤10 µm diameter) beams have already been shown to be essential. Nothing can replace an in-depth knowledge of crystals of a particular sample, but a metric of the extent of variability could be useful in providing an early indication of the level of of samples for a particular project.
|
In order to provide a general definition of sample variability, a normalized value should be assigned in order to be comparable between samples and projects. Two measures present themselves: the variance divided by the square of the mean (1) and the peak value divided by the mean (2),
where Tk is the TIS of a position within a crystal, is the average TIS, Tmax is the highest value of the TIS and n is the number of positions. For perfectly homogenous crystals V1 = 0 as there will be no variation about the mean and V2 = 1 as the peak will be equal to the mean.
Mesh and line scans were performed on a variety of samples and values of V1 and V2 were calculated (Table 1). The values were calculated for separate crystals, except for RhoA, where each value was calculated from a different orientation of the same crystal (ω = 0, 30, 60, 90 and 120°). The values obtained vary considerably and, from knowledge of the samples studied, provide a good measure of the degree to which the diffraction quality varies within a crystal (Fig. 1). The values are highly correlated (R2 = 0.98) and most crystals examined here have V1 < 2 and V2 < 4, which indicate a high degree of Could the variation in the calculated values arise from factors other than heterogeneity? An explanation for greater variability within some crystals could be that crystal size affects variability, either through increasing disorder or simply by an increase in the number of positions sampled. Plotting V1 against crystal size shows the value does not necessarily increase with increasing crystal size (Fig. 2). As a further control, values of V1 and V2 were calculated for 30° angular ranges collected from single positions within the crystals with the highest observed values (F1-ATPase and RhoA). The values obtained from these data sets (V1 = 0.02 for both systems) show that the intensity of the diffraction images varies very little, even when collected over a rotation range (Table 1). This demonstrates that the values reasonably reflect the difference in diffraction quality between positions within a crystal.
2.4. Model calculations
Scanning a beam across a crystal yields a patchwork of TIS values corresponding to local variations in the quality of the crystal. As the variance has been normalized, the values of both V1 and V2 are characteristic only of the crystal and its orientation; these measures are independent of the beam intensity and allow comparison of different orientations and indeed different crystals. A large variance implies high local variability in crystal quality. Further insight can be gained by employing a model construct which, although limited, is nonetheless illuminating. If the variability of a given specimen were precisely known, V1 and V2 could be calculated (as could any other measure). As it is, these two measures can determine two numbers more obviously related to crystal variability, but only two. We suppose that each position in the mesh belongs to one of two classes. The lower class yields a TIS of magnitude 1 and the other class a magnitude h > 1. We suppose that the ratio of the number of positions of magnitude 1 to those of magnitude h is N. Thus, a crystal diffracting strongly from one local area will have a large value of h and a large value of N. From any given data set, the values of h and N in the simple equivalent model can be constructed. Given h and N, the values of V1 and V2 can be calculated; such an imaginary sample would yield (3) and (4),
These equations can be solved to yield the parameters h and N in terms of measured values of V1 and V2 (equations 5 and 6),
Finally, either h or N can be eliminated, to give, for example, a relation between V1 and V2. If h is eliminated, the relation between V1 and V2 defines a curve for constant N, a measure of the variegation of the crystal (7),
Plots of values of V1 and V2 for ratios N of between 1 and 64 for increasing values of h are shown in Fig. 1. In different orientations the same crystal will approximately define such a curve when scanned in different orientations, as for RhoA, where the ratio of good regions to bad remains constant but h changes. The model curves provide an indication of how variation within a crystal is reflected in the values of V1 and V2. The most ordered systems fall below values that represent a ratio N of 10:1 (Fig. 1, red line). Crystals with higher variation (GPCR, pyrophosphatase, F1-ATPase and RhoA) have ratios N equivalent in the model to between 15:1 and 25:1 accompanied by large differences in diffraction intensity (h). The most striking is the small ribosome crystal, where a very large ratio is observed. The model demonstrates that the combination of both values is essential, as when a small number of good areas are present low V1 can mask a high V2.
Do the values relate to isomorphism between positions? The purpose of these values is to give an idea of the variation in intensity within a crystal in order to assist in data-collection strategy, rather than to determine whether areas will be isomorphous. However, in order to determine if there is a relationship, correlation coefficients of scaled intensities [CCI(i, j) values output by XSCALE; Kabsch, 2010] between data sets collected at different positions for crystals with low and high ratios N were examined. Four data sets were collected from different positions of trypsin (N = 2.1) and F1-ATPase (N = 14.1) crystals. For trypsin, all data sets had correlation coefficients above 0.99, indicating a high degree of isomorphism. The F1-ATPase data sets showed different behaviour. Data sets were collected from three high-intensity positions and one other position. Correlation coefficients between these positions varied from 0.26 between low-intensity and high-intensity positions to 0.96–0.99 for high-intensity positions. While this is a rather limited investigation, it demonstrates that data from crystals with a low ratio N can be merged better than those with a high ratio N. It is also clear that while diffraction varies considerably between positions in F1-ATPase crystals, some are nevertheless isomorphous. Clustering techniques could then be used to group isomorphous data sets from multiple positions (Foadi et al., 2013; Giordano et al., 2012).
3. Discussion
The ability to measure the extent of crystal variability is an important step in quantifying empirical intra-crystal and inter-crystal variability. As can be seen in Table 1, the values vary greatly between systems, with the most ordered systems studied here maintaining a ratio N < 10 and V1 < 2. Obtaining values below this cutoff suggests data-collection strategies that use the full crystal volume, either by using a large beam, translating the crystal during data collection or merging multiple data sets. In the most extreme examples shown here, F1-ATPase and the small G protein RhoA, the crystals have high values of both V1 and V2, the latter showing that parts of the crystals have an intensity ∼15 and ∼11 times higher than the mean, respectively (Table 1 and Fig. 1). How can these values be used to guide data-collection strategies for these and subsequent crystals? Take F1-ATPase first. Scanning the crystal with a 7 µm diameter X-ray beam reveals high variability in diffraction intensity, with no contiguous regions. The subsequent strategy will depend on the purpose of the experiment, but the use of a large beam and/or a helical strategy can be ruled out. While RhoA crystals also have high V values, the high-intensity points form a contiguous region approximately 100 µm in diameter. This implies that samples for this project should be scanned to locate the best region, but that a beam diameter of 100 µm would make better use of the diffraction volume, either by changing the beam focus, shaping or using a helical strategy across the optimal volume.
The values obtained also support data-collection strategies that are already used in highly challenging projects. For example, very low values are obtained for a large ribosome crystals (N = 5.3), advocating data collection using as much of the crystal volume as possible, the strategy that was used by the Ramakrishnan laboratory when collecting data from these crystals (Clemons et al., 2001), often defocusing the beam. Of particular interest are samples that have a low value of V1 but a rather high value of V2, most notably GPCR and pyrophosphatase TVP crystals, both of which are membrane proteins. This situation arises when a crystal contains a very few hot spots that are much higher than the average. Most GPCR crystals have low V values as the intensity is poor throughout the crystals; however, some crystals contain 1–3 regions with much higher intensity. This reflects the experiences of data collection using these crystals, where a large number (often in the thousands) need to be screened and only certain volumes diffract sufficiently using a microfocused beam (Warne et al., 2008). This is also the case with crystals of the membrane-bound pyrophosphatase, where using a microfocused beam was essential to collect data from the most ordered regions of a crystal (Kellosalo et al., 2012). Misleading values are obtained in cases where multiple lattices are present, where low values are obtained as the average intensity is very similar. Here, combination with spot counting or indexing could be used to flag arising from multiple lattices. The possibility of estimating the B factor of images would provide an excellent measure of quality that does not possess the problems associated with TIS.
With some care, the value could also be applied between crystals. In this study, separate crystals of insulin and glucose isomerase showed a high level of V1 and V2 between crystals, respectively). This could be extended to multiple crystals on separate supports, but care would have to be taken in order to collect characterization images using the same resolution and beam intensity. In this way, a measure of the variability between samples could be obtained. With the advent of screening the diffraction properties of crystals in crystallization plates (Jacquamet et al., 2009; Axford et al., 2012), the value could provide an indicator of the level of screening that will be required for a particular project.
where multiple microcrystals were present on the same support (values of 0.25 and 1.83 for insulin and 0.93 and 3.12 for glucose isomerase forWith the foreseen increase in the automation of sample evaluation, the values presented here could become one of the measures of quality associated with a sample. The number of samples evaluated at the ESRF and other facilities has been steadily increasing (Bowler et al., 2010), and the advent of pixel-array detectors has enabled many more positions within crystals to be evaluated by reducing data-collection times (Broennimann et al., 2006; Aishima et al., 2010). Sample evaluation on this scale will require additional measures of crystal quality, such as sample variability. The values could then be used, in combination with current indicators of crystal quality, to assist in sample selection and choice of beamline and data-collection strategy. The measures are now calculated for all mesh scans performed on the ESRF Structural Biology beamlines and are stored in the LIMS system ISPyB (Delagenière et al., 2011). The values determined so far are shown in Fig. 3.
Acknowledgements
We thank Olof Svensson (ESRF, Grenoble) for implementing the calculations of variability and determining the size of homogenous diffraction areas in the mesh-scan workflows. We also thank John Walker and Martin Montgomery (MRC Mitochondrial Biology Unit, Cambridge, England) for purified F1-ATPase. We are grateful to the Partnership for Structural Biology (PSB), Grenoble for an integrated structural biology environment.
References
Aishima, J., Owen, R. L., Axford, D., Shepherd, E., Winter, G., Levik, K., Gibbons, P., Ashton, A. & Evans, G. (2010). Acta Cryst. D66, 1032–1035. Web of Science CrossRef CAS IUCr Journals Google Scholar
Axford, D. et al. (2012). Acta Cryst. D68, 592–600. Web of Science CrossRef CAS IUCr Journals Google Scholar
Baxter, N. J., Bowler, M. W., Alizadeh, T., Cliff, M. J., Hounslow, A. M., Wu, B., Berkowitz, D. B., Williams, N. H., Blackburn, G. M. & Waltho, J. P. (2010). Proc. Natl Acad. Sci. USA, 107, 4555–4560. Web of Science CrossRef CAS PubMed Google Scholar
Bourenkov, G. P. & Popov, A. N. (2010). Acta Cryst. D66, 409–419. Web of Science CrossRef CAS IUCr Journals Google Scholar
Bowler, M. W., Guijarro, M., Petitdemange, S., Baker, I., Svensson, O., Burghammer, M., Mueller-Dieckmann, C., Gordon, E. J., Flot, D., McSweeney, S. M. & Leonard, G. A. (2010). Acta Cryst. D66, 855–864. Web of Science CrossRef CAS IUCr Journals Google Scholar
Bowler, M. W., Montgomery, M. G., Leslie, A. G. W. & Walker, J. E. (2006). Acta Cryst. D62, 991–995. Web of Science CrossRef CAS IUCr Journals Google Scholar
Brockhauser, S., Svensson, O., Bowler, M. W., Nanao, M., Gordon, E., Leal, R. M. F., Popov, A., Gerring, M., McCarthy, A. A. & Gotz, A. (2012). Acta Cryst. D68, 975–984. Web of Science CrossRef CAS IUCr Journals Google Scholar
Broennimann, C., Eikenberry, E. F., Henrich, B., Horisberger, R., Huelsen, G., Pohl, E., Schmitt, B., Schulze-Briese, C., Suzuki, M., Tomizaki, T., Toyokawa, H. & Wagner, A. (2006). J. Synchrotron Rad. 13, 120–130. Web of Science CrossRef CAS IUCr Journals Google Scholar
Cherezov, V., Hanson, M. A., Griffith, M. T., Hilgart, M. C., Sanishvili, R., Nagarajan, V., Stepanov, S., Fischetti, R. F., Kuhn, P. & Stevens, R. C. (2009). J. R. Soc. Interface, 6, S587–S597. Web of Science CrossRef PubMed CAS Google Scholar
Clemons, W. M. Jr, Brodersen, D. E., McCutcheon, J. P., May, J. L., Carter, A. P., Morgan-Warren, R. J., Wimberly, B. T. & Ramakrishnan, V. (2001). J. Mol. Biol. 310, 827–843. Web of Science CrossRef PubMed CAS Google Scholar
Cliff, M. J., Bowler, M. W., Varga, A., Marston, J. P., Szabó, J., Hounslow, A. M., Baxter, N. J., Blackburn, G. M., Vas, M. & Waltho, J. P. (2010). J. Am. Chem. Soc. 132, 6507–6516. Web of Science CrossRef CAS PubMed Google Scholar
Cusack, S., Belrhali, H., Bram, A., Burghammer, M., Perrakis, A. & Riekel, C. (1998). Nature Struct. Biol. 5, 634–637. Web of Science CrossRef CAS PubMed Google Scholar
Delagenière, S. et al. (2011). Bioinformatics, 27, 3186–3192. Web of Science PubMed Google Scholar
Efremov, R. G., Baradaran, R. & Sazanov, L. A. (2010). Nature (London), 465, 441–445. Web of Science CrossRef CAS PubMed Google Scholar
Flot, D., Mairs, T., Giraud, T., Guijarro, M., Lesourd, M., Rey, V., van Brussel, D., Morawe, C., Borel, C., Hignette, O., Chavanne, J., Nurizzo, D., McSweeney, S. & Mitchell, E. (2010). J. Synchrotron Rad. 17, 107–118. Web of Science CrossRef CAS IUCr Journals Google Scholar
Foadi, J., Aller, P., Alguel, Y., Cameron, A., Axford, D., Owen, R. L., Armour, W., Waterman, D. G., Iwata, S. & Evans, G. (2013). Acta Cryst. D69, 1617–1632. Web of Science CrossRef CAS IUCr Journals Google Scholar
Gabadinho, J. et al. (2010). J. Synchrotron Rad. 17, 700–707. Web of Science CrossRef CAS IUCr Journals Google Scholar
Giordano, R., Leal, R. M. F., Bourenkov, G. P., McSweeney, S. & Popov, A. N. (2012). Acta Cryst. D68, 649–658. Web of Science CrossRef CAS IUCr Journals Google Scholar
Griffin, J. L., Bowler, M. W., Baxter, N. J., Leigh, K. N., Dannatt, H. R., Hounslow, A. M., Blackburn, G. M., Webster, C. E., Cliff, M. J. & Waltho, J. P. (2012). Proc. Natl Acad. Sci. USA, 109, 6910–6915. Web of Science CrossRef CAS PubMed Google Scholar
Hilgart, M. C., Sanishvili, R., Ogata, C. M., Becker, M., Venugopalan, N., Stepanov, S., Makarov, O., Smith, J. L. & Fischetti, R. F. (2011). J. Synchrotron Rad. 18, 717–722. Web of Science CrossRef CAS IUCr Journals Google Scholar
Incardona, M.-F., Bourenkov, G. P., Levik, K., Pieritz, R. A., Popov, A. N. & Svensson, O. (2009). J. Synchrotron Rad. 16, 872–879. Web of Science CrossRef IUCr Journals Google Scholar
Jacquamet, L., Joly, J., Bertoni, A., Charrault, P., Pirocchi, M., Vernede, X., Bouis, F., Borel, F., Périn, J.-P., Denis, T., Rechatin, J.-L. & Ferrer, J.-L. (2009). J. Synchrotron Rad. 16, 14–21. Web of Science CrossRef CAS IUCr Journals Google Scholar
Kabsch, W. (2010). Acta Cryst. D66, 125–132. Web of Science CrossRef CAS IUCr Journals Google Scholar
Kellosalo, J., Kajander, T., Kogan, K., Pokharel, K. & Goldman, A. (2012). Science, 337, 473–476. Web of Science CrossRef CAS PubMed Google Scholar
Lutter, R., Abrahams, J. P., van Raaij, M. J., Todd, R. J., Lundqvist, T., Buchanan, S. K., Leslie, A. G. W. & Walker, J. E. (1993). J. Mol. Biol. 229, 787–790. CrossRef CAS PubMed Web of Science Google Scholar
Mueller-Dieckmann, C., Panjikar, S., Tucker, P. A. & Weiss, M. S. (2005). Acta Cryst. D61, 1263–1272. Web of Science CrossRef CAS IUCr Journals Google Scholar
Pellegrini, E., Piano, D. & Bowler, M. W. (2011). Acta Cryst. D67, 902–906. Web of Science CrossRef CAS IUCr Journals Google Scholar
Powell, H. R., Johnson, O. & Leslie, A. G. W. (2013). Acta Cryst. D69, 1195–1203. Web of Science CrossRef CAS IUCr Journals Google Scholar
Riekel, C., Burghammer, M. & Schertler, G. (2005). Curr. Opin. Struct. Biol. 15, 556–562. Web of Science CrossRef PubMed CAS Google Scholar
Russi, S., Juers, D. H., Sanchez-Weatherby, J., Pellegrini, E., Mossou, E., Forsyth, V. T., Huet, J., Gobbo, A., Felisaz, F., Moya, R., McSweeney, S. M., Cusack, S., Cipriani, F. & Bowler, M. W. (2011). J. Struct. Biol. 175, 236–243. Web of Science CrossRef CAS PubMed Google Scholar
Sanchez-Weatherby, J., Bowler, M. W., Huet, J., Gobbo, A., Felisaz, F., Lavault, B., Moya, R., Kadlec, J., Ravelli, R. B. G. & Cipriani, F. (2009). Acta Cryst. D65, 1237–1246. Web of Science CrossRef CAS IUCr Journals Google Scholar
Sanishvili, R., Nagarajan, V., Yoder, D., Becker, M., Xu, S., Corcoran, S., Akey, D. L., Smith, J. L. & Fischetti, R. F. (2008). Acta Cryst. D64, 425–435. Web of Science CrossRef CAS IUCr Journals Google Scholar
Sauter, N. K., Grosse-Kunstleve, R. W. & Adams, P. D. (2004). J. Appl. Cryst. 37, 399–409. Web of Science CrossRef CAS IUCr Journals Google Scholar
Selmer, M., Dunham, C. M., Murphy, F. V. IV, Weixlbaumer, A., Petry, S., Kelley, A. C., Weir, J. R. & Ramakrishnan, V. (2006). Science, 313, 1935–1942. Web of Science CrossRef PubMed CAS Google Scholar
Smith, J. L., Fischetti, R. F. & Yamamoto, M. (2012). Curr. Opin. Struct. Biol. 22, 602–612. Web of Science CrossRef CAS PubMed Google Scholar
Song, J., Mathew, D., Jacob, S. A., Corbett, L., Moorhead, P. & Soltis, S. M. (2007). J. Synchrotron Rad. 14, 191–195. Web of Science CrossRef CAS IUCr Journals Google Scholar
Warne, T., Serrano-Vega, M. J., Baker, J. G., Moukhametzianov, R., Edwards, P. C., Henderson, R., Leslie, A. G. W., Tate, C. G. & Schertler, G. F. (2008). Nature (London), 454, 486–491. Web of Science CrossRef PubMed CAS Google Scholar
Wheeler, M. J., Russi, S., Bowler, M. G. & Bowler, M. W. (2012). Acta Cryst. F68, 111–114. Web of Science CrossRef CAS IUCr Journals Google Scholar
Zhang, Z., Sauter, N. K., van den Bedem, H., Snell, G. & Deacon, A. M. (2006). J. Appl. Cryst. 39, 112–119. Web of Science CrossRef CAS IUCr Journals Google Scholar