Computer-controlled liquid-nitrogen drizzling device for removing frost from cryopreserved crystals

CROSSMARK_Color_square_no_text.svg

aProtein Crystal Analysis Division, Japan Synchrotron Radiation Research Institute, 1-1-1 Kouto, Sayo-cho, Sayo-gun, Hyogo 679-5198, Japan, bAdvanced Photon Technology Division, RIKEN SPring-8 Center, 1-1-1 Kouto, Sayo-cho, Sayo-gun, Hyogo 679-5148, Japan, and cROD (Single Crystal Analysis) Group, Application Laboratories, Rigaku Corporation, 3-9-11 Matsubara-cho, Akishima-shi, Tokyo 196-8666, Japan
*Correspondence e-mail: kumasaka@spring8.or.jp

Edited by J. Newman, Bio21 Collaborative Crystallisation Centre, Australia (Received 8 October 2020; accepted 24 October 2020; online 25 November 2020)

Cryocrystallography is a technique that is used more often than room-temperature data collection in macromolecular crystallography. One of its advantages is the significant reduction in radiation damage, which is especially useful in synchrotron experiments. Another advantage is that cryopreservation provides simple storage of crystals and easy transportation to a synchrotron. However, this technique sometimes results in the undesirable adhesion of frost to mounted crystals. The frost produces noisy diffraction images and reduces the optical visibility of crystals, which is crucial for aligning the crystal position with the incident X-ray position. To resolve these issues, a computer-controlled device has been developed that drizzles liquid nitrogen over a crystal to remove frost. It was confirmed that the device works properly, reduces noise from ice rings in diffraction images and enables the centering of crystals with low visibility owing to frost adhesion.

1. Introduction

In macromolecular crystallography (MX) using a synchrotron, cryocrystallography is commonly used to obtain high-quality diffraction data (Hope, 1988[Hope, H. (1988). Acta Cryst. B44, 22-26.]; Parkin & Hope, 1998[Parkin, S. & Hope, H. (1998). J. Appl. Cryst. 31, 945-953.]; Rodgers, 1994[Rodgers, D. W. (1994). Structure, 2, 1135-1140.]; Garman & Schneider, 1997[Garman, E. F. & Schneider, T. R. (1997). J. Appl. Cryst. 30, 211-237.]). The sample temperature is maintained at 100 K or below using a cryostream device and is important for reducing radiation damage to the samples and automating sample exchange (Murakami et al., 2012[Murakami, H., Ueno, G., Shimizu, N., Kumasaka, T. & Yamamoto, M. (2012). J. Appl. Cryst. 45, 234-238.]).

However, a major obstacle in this technique is ice or frost contamination in or on the flash-cooled droplets containing crystals. This deteriorates the diffraction images by the addition of reflections derived from ice diffraction. Frost adhesion and devitrification of solvent ice also lower the optical visibility of crystals, thus interfering with the process of aligning the crystal position with the incident X-ray position.

Ice formation is basically resolved by optimizing the harvesting solvent and/or cooling rate during the cryo­preservation process (Harrison et al., 2019[Harrison, K., Wu, Z. & Juers, D. H. (2019). J. Appl. Cryst. 52, 1222-1232.]; Warkentin et al., 2006[Warkentin, M., Berejnov, V., Husseini, N. S. & Thorne, R. E. (2006). J. Appl. Cryst. 39, 805-811.]), but the problem of frost mainly comes from the sample-handling environment. High humidity causes the condensation of water vapor contained in the air on cooled materials, such as samples, the sample changer, goniometer head and cryocooler nozzle. Even though this can be avoided by heating the devices (Russi et al., 2016[Russi, S., Song, J., McPhillips, S. E. & Cohen, A. E. (2016). J. Appl. Cryst. 49, 622-626.]) or using dry air such as a sheath gas stream in a cryostream, it is inevitable in some cases, i.e. during transportation of the samples using dry shippers. Such frost is formed in the atmosphere and then adheres to the sample; therefore, frost removal is possible in many cases.

There are methods for removing frost from crystals mounted on goniometers. One is annealing (Harp et al., 1998[Harp, J. M., Timm, D. E. & Bunick, G. J. (1998). Acta Cryst. D54, 622-628.]; Yeh & Hol, 1998[Yeh, J. I. & Hol, W. G. J. (1998). Acta Cryst. D54, 479-480.]), which is mostly used to improve crystal quality by removing the crystal disorder that is introduced during fast quenching (Kriminski et al., 2002[Kriminski, S., Caylor, C. L., Nonato, M. C., Finkelstein, K. D. & Thorne, R. E. (2002). Acta Cryst. D58, 459-471.]). To increase the temperature of crystals, blocking of the cold stream is suitable for automatic operation (Yeh & Hol, 1998[Yeh, J. I. & Hol, W. G. J. (1998). Acta Cryst. D54, 479-480.]) and is applied at the SPring-8 MX beamlines. This defrosting method can effectively remove frost, but some fragile samples may be damaged by the temperature change. There is also a method that involves rinsing the sample pins in the liquid nitrogen (LN2) dewar with a sample-handling robot; however, there is the possibility of frost adhering again (Soltis et al., 2008[Soltis, S. M., Cohen, A. E., Deacon, A., Eriksson, T., González, A., McPhillips, S., Chui, H., Dunten, P., Hollenbeck, M., Mathews, I., Miller, M., Moorhead, P., Phizackerley, R. P., Smith, C., Song, J., van dem Bedem, H., Ellis, P., Kuhn, P., McPhillips, T., Sauter, N., Sharp, K., Tsyba, I. & Wolf, G. (2008). Acta Cryst. D64, 1210-1221.]). Another method involves flashing a refrigerant medium such as liquid nitrogen over the crystal (Pflugrath, 2004[Pflugrath, J. W. (2004). Methods, 34, 415-423.]; Warkentin & Thorne, 2007[Warkentin, M. & Thorne, R. E. (2007). J. Struct. Funct. Genomics, 8, 141-144.]; Garman & Owen, 2006[Garman, E. F. & Owen, R. L. (2006). Acta Cryst. D62, 32-47.]). This method can remove frost and dust from a crystal by liquid flow. Even though some of the fixed dust still remains with certain flow rates and amounts of liquid, a sample can maintain a low temperature during the process.

The flashing-of-refrigerant method is advantageous for any sample. However, there are issues with its implementation on synchrotron beamlines. Recent MX beamlines are highly automated and many probes and instruments are densely packed around the sample position. In this situation, LN2-flashing devices must be compact and prevent the spillage of LN2 onto the other devices. In addition, computer-controlled operation is required because automation and remote measurements are widely carried out (Ueno et al., 2006[Ueno, G., Kanda, H., Hirose, R., Ida, K., Kumasaka, T. & Yamamoto, M. (2006). J. Struct. Funct. Genomics, 7, 15-22.], 2016[Ueno, G., Hikima, T., Yamashita, K., Hirata, K., Hasegawa, K., Murakami, H., Furukawa, Y., Mizuno, N., Kumasaka, T. & Yamamoto, M. (2016). AIP Conf. Proc. 1741, 050021.]; Hirata et al., 2019[Hirata, K., Yamashita, K., Ueno, G., Kawano, Y., Hasegawa, K., Kumasaka, T. & Yamamoto, M. (2019). Acta Cryst. D75, 138-150.]). Thus, the automatic sequence of setting up data collection must be carried out without human intervention and users must be able to apply the LN2 to the sample when they are controlling the experiment remotely. To solve these problems, we developed a computer-controlled device that automatically drizzles LN2. We describe its configuration and the results of diffraction experiments with and without the device.

2. Device and usage

2.1. Device configuration

The device consists of four major components, as shown in Fig. 1[link](a). The dewar tank is used to store LN2 and is equipped with a heater unit and a solenoid valve. To blow out LN2, the heater is turned on and the valve is closed by the increasing gas pressure inside the dewar (Fig. 1[link]b). The dewar also has upper- and lower-limit fluid-level sensors that are used when refilling the LN2. The LN2 is blown out through an insulated transfer tube and is transferred to an LN2–gas separator. The blowout contains nitrogen gas, which causes intermittent LN2 flow. Therefore, the nitrogen gas should be discarded using the separator, which is made of a cylindrical aluminium tube with a filter plate with four 0.7 mm diameter holes (Fig. 1[link]c). Only LN2 flows through these holes and passes down to the next LN2 drizzling nozzle, and the gas flowing out from the LN2 passes through the top hole of the separator tube (Fig. 1[link]d). The gas is then completely removed by the funnel-shaped top of the LN2-drizzling nozzle (5 mm internal diameter) and LN2 is delivered to the sample position. The distance from the tip of the LN2-drizzling nozzle to the sample position is 15 mm.

[Figure 1]
Figure 1
(a) The device at BL41XU. (b) A configuration diagram of the device. (c) A drawing of the LN2–gas separator. (d) The LN2–gas separator at BL26B2 (before installing the insulation tube). (e) A close-up view of the device at the sample position (the device is shown in pink).

This device is designed to be compact for installation without interfering with the sample position, diffractometer and other beamline devices such as the cryostream nozzle, beam stopper and collimator (Fig. 1[link]e).

In the development of this device, it was important to avoid blowing nitrogen gas, which can cause a temperature increase at the sample position. As mentioned previously, only LN2 is able to drizzle onto the sample by completely relieving pressure using the two stages of the LN2–gas separator and LN2-drizzling nozzle. The LN2 supply parts (dewar tank, heater unit, solenoid valve and level sensors) were procured from Taiyo Nippon Sanso Corp., Tokyo, Japan. The LN2–gas separator and LN2-drizzling nozzle are both standard metalwork products and cost a total of approximately 1000 dollars.

2.2. Applications in various measurement situations

The device can be operated via computers installed at the SPring-8 MX beamlines. For on-demand operation, both on-site and remote users can operate it through the control GUIs provided by the MX beamline-control software BSS (Ueno et al., 2005[Ueno, G., Kanda, H., Kumasaka, T. & Yamamoto, M. (2005). J. Synchrotron Rad. 12, 380-384.]) and the remote-measurement software SP8Remote (Ueno et al., 2016[Ueno, G., Hikima, T., Yamashita, K., Hirata, K., Hasegawa, K., Murakami, H., Furukawa, Y., Mizuno, N., Kumasaka, T. & Yamamoto, M. (2016). AIP Conf. Proc. 1741, 050021.]). Both interfaces simply provide one-button operation.

The device has also been integrated into automatic measurement systems. DeepCentering is a novel system which performs automatic crystal recognition and centering by using deep-learning technology (Ito et al., 2019[Ito, S., Ueno, G. & Yamamoto, M. (2019). J. Synchrotron Rad. 26, 1361-1366.]). This system is designed to analyze clean sample photographs; however, frost adhesion to the sample loop interferes with the correct extraction of the outline of the crystals. For this reason, LN2 is drizzled in advance using the device before taking a crystal photograph. Its operation is incorporated into the automatic centering protocol of this system.

The ZOO system for automatic diffraction measurements (Hirata et al., 2013[Hirata, K., Kawano, Y., Ueno, G., Hashimoto, K., Murakami, H., Hasegawa, K., Hikima, T., Kumasaka, T. & Yamamoto, M. (2013). J. Phys. Conf. Ser. 425, 012002.]) also supports the device. In this system, crystal centering is performed using both optical image recognition and X-ray scanning. For image recognition, the loop is centered and oriented using the following steps: (i) the outline shape of the loop is recognized from the sample photograph and the center of the loop is adjusted to the X-ray position and (ii) the goniometer φ angle is determined to orthogonalize the loop plane against the X-ray beam axis by maximizing the projected area of the loop (facing). Facing is especially important for small wedge measurements, in which each microcrystal in the area is serially measured over a small φ-angle range. In the case of adhered frost, the outer shape of the nylon loop will not be detected accurately. Additionally, low accuracy in facing, i.e. the inclination of the plane of the loop, increases the path length of the X-rays through the sample, thus increasing the rate of multi-hit diffraction images, which are not suitable for data processing. Additionally, adhered frost results in noisy ice diffraction during data-set collection, which interferes with accurate measurements.

3. Results and discussion

3.1. Device operation and visual inspection of crystals before and after drizzling LN2

The device can continuously transport an LN2 flow to the sample position at a flow rate of 0.4–0.5 ml s−1. The rate varies depending on the amount of LN2 remaining in the tank, but it does not severely change the ability to remove frost. In our experience, one-time drizzling for about 10 s is sufficient to clean the crystals. This device can remove ice and fibrous dust adhering to the surface. However, it sometimes failed to completely remove very fine frost (Fig. 2[link]f). This type of frost, powder-like frost with dimensions of approximately a few micrometres, often occurs during long-term storage in dry shippers and might also occur at low temperatures and lower supersaturation conditions according to the Nakaya snow-crystal morphology diagram (Nakaya, 1954[Nakaya, U. (1954). Snow Crystals: Natural and Artificial. Cambridge: Harvard University Press.]). In such a case, longer drizzling can be applied. However, the nozzle sometimes blocks up owing to the formation of frost on it when LN2 is drizzled continuously for more than 180 s.

[Figure 2]
Figure 2
Examples of frost removal from samples. (1) Before and (2) after LN2 drizzling. Samples (af) are crystals from beamline users. In sample (e) fibrous dust was removed and in sample (f) very fine frost was not completely removed. Sample (g) is a lysozyme crystal and image (g), 2 shows a dark shadow caused by X-ray exposure. Sample (h) is the lysozyme crystal described in Section 3.3[link].

Examples of sample-crystal photographs are shown in Fig. 2[link] and were taken using sample-observation CCD cameras installed on the beamlines. After drizzling, the adhered frost was completely removed and the crystal visibility improved. We confirmed that not only frost but also unidentified ice or dust was removed. The temperature variation of the samples owing to the drizzling was measured using a simple method involving a type-K thermocouple; the temperature dropped to 86 K from 127 K within 0.25 s after LN2 was drizzled.

3.2. Time of frost removal

We compared the times required for frost removal by drizzling LN2 using the device and by using a manual LN2-drizzling tool (Pflugrath, 2004[Pflugrath, J. W. (2004). Methods, 34, 415-423.]; Warkentin & Thorne, 2007[Warkentin, M. & Thorne, R. E. (2007). J. Struct. Funct. Genomics, 8, 141-144.]). The experimental protocol was assumed to be a standard experiment at a SPring-8 MX beamline; that is, it includes automated sample exchange using a SPACE robot (Murakami et al., 2012[Murakami, H., Ueno, G., Shimizu, N., Kumasaka, T. & Yamamoto, M. (2012). J. Appl. Cryst. 45, 234-238.]). The time was measured from immediately after setting the sample pin on the goniometer to starting data collection. In both cases, LN2 was drizzled for 10 s. The average time for each of the five samples was 106.0 ± 4.5 s for manual operation and 66.8 ± 5.8 s for the device. This shows that the device is about 40 s faster per sample. For manual operation, the time included an operator opening/closing the door of the experimental hutch to drizzle LN2, which took about 20 s.

3.3. Comparison of diffraction data before and after frost removal

To evaluate the effect of frost adhesion in data collection, diffraction data were collected from the same crystal before and after frost removal using the device. The measurements were conducted under the same conditions as those reported for beamline BL26B2 at SPring-8 (Ueno et al., 2006[Ueno, G., Kanda, H., Hirose, R., Ida, K., Kumasaka, T. & Yamamoto, M. (2006). J. Struct. Funct. Genomics, 7, 15-22.]; Table 1[link]). We used the tetragonal hen egg-white lysozyme crystal shown in Fig. 2[link](h).

Table 1
Measurement conditions

Beamline BL26B2, SPring-8
Detector Rayonix MX225HS
Wavelength (Å) 1.0000
Rotation range per image (°) 0.2
Total rotation range (°) 90.0
Exposure time per image (s) 0.2
Crystal-to-detector distance (mm) 110
Beam size (µm) 50 × 60
Beam shape Gaussian
Photon flux (photons s−1) 1.52 × 1010
Maximum dose (MGy) 0.425
†The absorbed dose was calculated using RADDOSE-3D (Zeldin et al., 2013[Zeldin, O. B., Gerstel, M. & Garman, E. F. (2013). J. Appl. Cryst. 46, 1225-1230.]).

The frost-adhered crystals were prepared using the following procedure: the crystals were picked up from a droplet with a cryoloop and held using cooled cryotongs (277 K). This was performed inside a cool workbench, which was operated at 277 K and located beside the beamline experimental hutch (Baba et al., 2019[Baba, S., Shimada, A., Mizuno, N., Baba, J., Ago, H., Yamamoto, M. & Kumasaka, T. (2019). J. Appl. Cryst. 52, 699-705.]). The crystal was transferred to the goniometer head and then flash-cooled at 100 K from the cryotongs. This cooling was carried out on the beamline at 308 K. The cooled cryotongs sometimes condense water vapor around the crystal during transportation.

The diffraction images were taken with the same oscillation range in both data sets, as shown in Fig. 3[link]. Debye rings of ice diffraction were initially observed at all oscillation angles, but they completely disappeared after LN2 drizzling. Both rings correspond to ice-crystal reflections, as shown in Fig. 3[link](c).

[Figure 3]
Figure 3
Diffraction images before (a) and after (b) frost removal. (c) The image in (a) with the Debye rings from ice crystals marked with the relevant resolutions (Malkin et al., 2012[Malkin, T. L., Murray, B. J., Brukhno, A. V., Anwar, J. & Salzmann, C. G. (2012). Proc. Natl Acad. Sci. USA, 109, 1041-1045.]; Chapman & Somasundaram, 2010[Chapman, M. S. & Somasundaram, T. (2010). Acta Cryst. D66, 741-744.]). The ice rings and diffraction spots disappeared after drizzling the crystal with LN2.

Each data set was processed using XDS (Kabsch, 1988[Kabsch, W. (1988). J. Appl. Cryst. 21, 67-72.]), and the number of rejections in 100 resolution shells was determined, as shown in Fig. 4[link]. In the shells relevant to ice diffraction, the statistics improved significantly. In addition, after removing the ice Rmerge changed from 3.8% to 3.6% and I/σ(I) changed from 17.66 to 18.25 in the overall statistics (Table 2[link]).

Table 2
Data-processing statistics before and after LN2 drizzling

Values in parentheses are for the outermost shell.

  Before drizzling LN2 After drizzling LN2
Space group P43212 P43212
Unit-cell parameters (Å) a = 78.83, c = 36.95 a = 78.83, c = 36.95
Resolution range (Å) 39.41–1.20 (1.27–1.20) 39.41–1.20 (1.27–1.20)
No. of reflections (observed) 239029 (22035) 240059 (22199)
No. of reflections (unique) 66958 (9985) 67041 (10036)
Completeness (%) 96.2 (88.6) 96.4 (89.1)
Rmerge (%) 3.8 (75.6) 3.6 (75.8)
CC1/2 (%) 100.0 (51.0) 100.0 (51.7)
I/σ(I)〉 17.66 (1.18) 18.25 (1.16)
Rmerge = [\textstyle \sum_{hkl}\sum_{i}|I_{i}(hkl)- \langle I(hkl)\rangle|/][\textstyle \sum_{hkl}\sum_{i}I_{i}(hkl)].
[Figure 4]
Figure 4
The number of rejected reflections in each resolution range. The difference in the applicable resolution ranges is extremely large. Asterisks indicate the resolutions marked in Fig. 3[link](c).

4. Conclusion

The LN2-drizzling device worked well to clean the crystal samples. Use of the device to remove surface frost has the following advantages: (i) there is no diffraction from the attached frost and/or dust, improving the experimental data, and (ii) the visibility in sample photographs is improved. It is also suggested that the device can contribute to the success rate of crystal centering during automatic measurements.

Furthermore, the computer-controlled automation of LN2 drizzling (i) improves experimental throughput by shortening the operation time of drizzling and (ii) enables the implementation of LN2 drizzling in remote experiments.

We started the development of this device in 2016, and it is currently available on SPring-8 MX beamlines BL26B2, BL41XU (Hasegawa et al., 2013[Hasegawa, K., Shimizu, N., Okumura, H., Mizuno, N., Baba, S., Hirata, K., Takeuchi, T., Yamazaki, H., Senba, Y., Ohashi, H., Yamamoto, M. & Kumasaka, T. (2013). J. Synchrotron Rad. 20, 910-913.]) and BL45XU.

Even though some types of frost may not be completely removed, as stated above, the challenge for the future is to apply more pressure in the LN2 dewar to increase the flow rate of LN2. For this purpose, it will be necessary to ensure the safety of pressurization, to prevent dew condensation in the flow path and to prevent nitrogen gas from being blown out of the drizzling nozzle.

Acknowledgements

All diffraction experiments were conducted on the SPring-8 MX beamlines with the approval of the Japan Synchrotron Radiation Research Institute (JASRI; proposal Nos. 2016B1970, 2017B2555, 2019A2548, 2019B2548 and 2019B8480). Author contributions were as follows. TK and SB supervised the project. YN and SB designed the device with the help of TI and Taiyo Nippon Sanso Corp. YN assembled, tested and improved the device. GU modified BSS for device control. KH, SI and YM modified the ZOO and Deep­Centering systems for device adoption. YN and TK wrote the manuscript and all others revised it.

Funding information

This work was partially supported by the Platform Project for Supporting Drug Discovery and Life Science Research (Basis for Supporting Innovative Drug Discovery and Life Science Research; BINDS) from AMED under Grant No. JP1.7am0101070.

References

First citationBaba, S., Shimada, A., Mizuno, N., Baba, J., Ago, H., Yamamoto, M. & Kumasaka, T. (2019). J. Appl. Cryst. 52, 699–705.  CrossRef CAS IUCr Journals Google Scholar
First citationChapman, M. S. & Somasundaram, T. (2010). Acta Cryst. D66, 741–744.  Web of Science CrossRef IUCr Journals Google Scholar
First citationGarman, E. F. & Owen, R. L. (2006). Acta Cryst. D62, 32–47.  Web of Science CrossRef CAS IUCr Journals Google Scholar
First citationGarman, E. F. & Schneider, T. R. (1997). J. Appl. Cryst. 30, 211–237.  CrossRef Web of Science IUCr Journals Google Scholar
First citationHarp, J. M., Timm, D. E. & Bunick, G. J. (1998). Acta Cryst. D54, 622–628.  Web of Science CrossRef CAS IUCr Journals Google Scholar
First citationHarrison, K., Wu, Z. & Juers, D. H. (2019). J. Appl. Cryst. 52, 1222–1232.  Web of Science CrossRef CAS IUCr Journals Google Scholar
First citationHasegawa, K., Shimizu, N., Okumura, H., Mizuno, N., Baba, S., Hirata, K., Takeuchi, T., Yamazaki, H., Senba, Y., Ohashi, H., Yamamoto, M. & Kumasaka, T. (2013). J. Synchrotron Rad. 20, 910–913.  Web of Science CrossRef CAS IUCr Journals Google Scholar
First citationHirata, K., Kawano, Y., Ueno, G., Hashimoto, K., Murakami, H., Hasegawa, K., Hikima, T., Kumasaka, T. & Yamamoto, M. (2013). J. Phys. Conf. Ser. 425, 012002.  CrossRef Google Scholar
First citationHirata, K., Yamashita, K., Ueno, G., Kawano, Y., Hasegawa, K., Kumasaka, T. & Yamamoto, M. (2019). Acta Cryst. D75, 138–150.  Web of Science CrossRef IUCr Journals Google Scholar
First citationHope, H. (1988). Acta Cryst. B44, 22–26.  CrossRef CAS Web of Science IUCr Journals Google Scholar
First citationIto, S., Ueno, G. & Yamamoto, M. (2019). J. Synchrotron Rad. 26, 1361–1366.  CrossRef CAS IUCr Journals Google Scholar
First citationKabsch, W. (1988). J. Appl. Cryst. 21, 67–72.  CrossRef CAS Web of Science IUCr Journals Google Scholar
First citationKriminski, S., Caylor, C. L., Nonato, M. C., Finkelstein, K. D. & Thorne, R. E. (2002). Acta Cryst. D58, 459–471.  Web of Science CrossRef CAS IUCr Journals Google Scholar
First citationMalkin, T. L., Murray, B. J., Brukhno, A. V., Anwar, J. & Salzmann, C. G. (2012). Proc. Natl Acad. Sci. USA, 109, 1041–1045.  Web of Science CrossRef CAS PubMed Google Scholar
First citationMurakami, H., Ueno, G., Shimizu, N., Kumasaka, T. & Yamamoto, M. (2012). J. Appl. Cryst. 45, 234–238.  Web of Science CrossRef CAS IUCr Journals Google Scholar
First citationNakaya, U. (1954). Snow Crystals: Natural and Artificial. Cambridge: Harvard University Press.  Google Scholar
First citationParkin, S. & Hope, H. (1998). J. Appl. Cryst. 31, 945–953.  Web of Science CrossRef CAS IUCr Journals Google Scholar
First citationPflugrath, J. W. (2004). Methods, 34, 415–423.  Web of Science CrossRef PubMed CAS Google Scholar
First citationRodgers, D. W. (1994). Structure, 2, 1135–1140.  CrossRef CAS PubMed Web of Science Google Scholar
First citationRussi, S., Song, J., McPhillips, S. E. & Cohen, A. E. (2016). J. Appl. Cryst. 49, 622–626.  Web of Science CrossRef CAS IUCr Journals Google Scholar
First citationSoltis, S. M., Cohen, A. E., Deacon, A., Eriksson, T., González, A., McPhillips, S., Chui, H., Dunten, P., Hollenbeck, M., Mathews, I., Miller, M., Moorhead, P., Phizackerley, R. P., Smith, C., Song, J., van dem Bedem, H., Ellis, P., Kuhn, P., McPhillips, T., Sauter, N., Sharp, K., Tsyba, I. & Wolf, G. (2008). Acta Cryst. D64, 1210–1221.  Web of Science CrossRef IUCr Journals Google Scholar
First citationUeno, G., Kanda, H., Hirose, R., Ida, K., Kumasaka, T. & Yamamoto, M. (2006). J. Struct. Funct. Genomics, 7, 15–22.  CrossRef PubMed CAS Google Scholar
First citationUeno, G., Kanda, H., Kumasaka, T. & Yamamoto, M. (2005). J. Synchrotron Rad. 12, 380–384.  Web of Science CrossRef IUCr Journals Google Scholar
First citationUeno, G., Hikima, T., Yamashita, K., Hirata, K., Hasegawa, K., Murakami, H., Furukawa, Y., Mizuno, N., Kumasaka, T. & Yamamoto, M. (2016). AIP Conf. Proc. 1741, 050021.  Google Scholar
First citationWarkentin, M., Berejnov, V., Husseini, N. S. & Thorne, R. E. (2006). J. Appl. Cryst. 39, 805–811.  Web of Science CrossRef CAS IUCr Journals Google Scholar
First citationWarkentin, M. & Thorne, R. E. (2007). J. Struct. Funct. Genomics, 8, 141–144.  CrossRef PubMed CAS Google Scholar
First citationYeh, J. I. & Hol, W. G. J. (1998). Acta Cryst. D54, 479–480.  Web of Science CrossRef CAS IUCr Journals Google Scholar
First citationZeldin, O. B., Gerstel, M. & Garman, E. F. (2013). J. Appl. Cryst. 46, 1225–1230.  Web of Science CrossRef CAS IUCr Journals Google Scholar

This is an open-access article distributed under the terms of the Creative Commons Attribution (CC-BY) Licence, which permits unrestricted use, distribution, and reproduction in any medium, provided the original authors and source are cited.

Journal logoSTRUCTURAL BIOLOGY
COMMUNICATIONS
ISSN: 2053-230X
Follow Acta Cryst. F
Sign up for e-alerts
Follow Acta Cryst. on Twitter
Follow us on facebook
Sign up for RSS feeds